[QUOTE=Napier]
Crafter_Man, you never fail to please.
[/QUOTE]
That’s what the women say. 
[QUOTE=Napier]
So, whether you can buy 3-wire IPRTs depends on where you draw the line between IPRT and RTD. I find this important semantic issue somewhat inscrutable. You can’t buy highly accurate 3-wire sensors. I actually haven’t found manufacturers that label their products IPRTs though I’m not far down that particular road.
[/QUOTE]
A few things:
- RTD simply means Resistance Temperature Device. There are many kinds of RTDs. Off the top of my head:
- 2-wire industrial platinum resistance thermometers (IPRTs)
- 3-wire IPRTs
- 4-wire IPRTs
- 4-wire standard platinum resistance thermometers (SPRTs)
- 2-wire industrial copper resistance thermometers (ICRTs)
- 3-wire ICRTs
- 2-wire industrial nickel resistance thermometers (INRTs)
- 3-wire INRTs
- PTC thermistors
- NTC thermistors
As can be seen, a 4-wire IPRTs is just one kind of RTD. But for whatever reason, everyone referes to a 4-wire IPRT simply as an RTD. While this is true, 4-wire IPRT is a more accurate description.
- The difference between a 3-wire IPRT and a 4-wire IPRT is simply that… one has 3 wires, and the other has 4 wires. The platinum elements are the same. Therefore, the intrinsic accuracies of each platinum element are the same. The reason the 3-wire IPRT is less accurate than 4-wire has to do with lead resistance compensation (as you have noted). With a 3-wire IPRT, the resistances of the leads will cancel as long as they’re matched fairly well, and they don’t change. Lead resistance doesn’t matter with a 4-wire IPRT, which is why they’re so much superior.
[QUOTE=Napier]
I think I’m trying to navigate from Omega RTD land, where equipment designers tend to congregate, towards PRT land but without losing the mechanically robust nature of RTDs in their wiring and connection hardware. Note that I understand the sensors themselves have robustness compromises with accuracy - hysteresis in particular - but the sensors are well protected and quiet; it’s the wiring and connectors I want to harden.
[/QUOTE]
Some RTDs are very mechanically robust (e.g. thermistors, film-type IPRTs), some are extremely delicate (e.g. SPRT), and some are in-between (e.g. IPRTs).
Here’s the deal: An IPRT is simply a platinum helix wrapped around a mica or ceramic insulator. Unlike SPRTs, an IPRT is not wrapped in a “strain-free” way. This allows the IPRT to survive a moderate amount of mechanical stress, but there is a price to pay: there is some hysteresis, and the repeatability is only so-so. But the uncertainty is still pretty impressive. If you thoroughly calibrate it, an IPRT can be assigned an uncertainty of around ± 0.05 °C over its entire operating range. But… the IPRT is *still * somewhat delicate. An IPRT *cannot * be smacked around or dropped on the floor, as doing so *will * throw R[sub]0[/sub] off.
We use SPRTs and IPRTs in our temperature metrology lab. Only two people are allowed to handle the SPRTs. We also have an IPRT that we take into the field for performing system temperature calibrations on certian things. To be safe, I instruct my field technician to handle the IPRT in much the same way she handles the SPRTs in the lab.
And you’re smart to think about wiring & connectors. They always seem to break. If you don’t care about size & aesthetics, look no further than military-style Amphenol connectors.
[QUOTE=Napier]
I think I also need better information about just how much accuracy we lose through 3-wire systems per se. I’m gonna guess it’s on the order of 0.05 or 0.1 C for clean new systems sitting on the benchtop, and maybe 0.3 or even sporadically 1 C or more for real systems with long wire bundles and infrequently remade connections that may go years without attention.
[/QUOTE]
It depends on a number of factors, but I would guess a 3-wire system adds an additional ± 0.3 °C or so. At any rate, if you want to use an IPRT, I would never even consider a 3-wire configuration. Go with a 4-wire configuration, and life will be much better.
[QUOTE=Napier]
Yes, I know about thermoelectric effects. Premade products I buy use reversing DC. When I program my own systems I randomize the current over a range like 0.5 to 1 mA and reverse the voltage and do the appropriate renormalization and averaging. Standard RTD systems I see in use now do neither of these things.
[/QUOTE]
Sounds good! And of course, you actively reverse the excitation current, and the voltage you’re measuring passively reverses.
And if you’re going to the trouble of addressing thermoelectric effects, then you must use a 4-wire IPRT. Finally, if you want uber-accuracy, measure the resistance at two different currents, and then calculate the “zero current resistance” using a second-order curve-fit. 
[QUOTE=Napier]
I guess the battles I have are those of wanting to do better than run-of-the-mill RTD systems, to get total error budgets to 0.x C given months or a couple years between servicing. All in the 0 to 500 C range in otherwise pleasant and noncorrosive and quiet settings.
[/QUOTE]
I have used IPRTs for years in our temperature metrology lab. Some things I have learned:
-
Treat the IPRT as if it were an SPRT. As an example, when you lay the IRPT on a table, there should not be a “tink” sound when its outer sheath touches the table. Yes, that is how I instruct my technicians to handle IPRTs! It is especially serious in the case of an SPRT. If one of my technicians lays an SPRT on a table, and I hear the sound of the SPRT’s stainless steel sheath touching the table, I immediately instruct them to stick the SPRT in the water triple point cell to see if the triple point resistance has shifted.
-
An IPRT must be calibrated at at least 5 points over its operating range at least once a year. The Callendar-Van Dusen equation is what I use to covert temperature to voltage. (I use the quartic solution to convert voltage to temperature for T < 0 °C.) In addition, I would suggest sticking the IPRT in an ice bath every now and then to see if its R[sub]0[/sub] has drifted.
-
An IRPT should be periodically annealed to relive molecular stresses in the platinum and to “boil off” impurities.